Is anyone using the green top Greiner lithium heparin (with gel) tubes for chemistry? Are you experiencing a problem with the gel not settling properly during centrifugation ? Gel seems to float in the plasma and causes huge issues with the chemistry analyzers.
-
They can range from 5.8 to over 6. The specimens are not refrigerated, not re-spun. Patients do not have I.V.'s as we are an outpatient facility and therefore no transfusions either. I've thought about draw order to the point we actually purposely drew 30 in the wrong order and all the K+'s were fine.
Thanks for the ...
-
We have spurious high postassiums at least 3-4 times per month. These samples usually are not hemolyzed, the patient is not on medication which could cause the problem. When drawn the tourniquet was not left on during the draw and the patient was not pumping their fist. When we redraw the patient we always get a normal ...
-
Can anyone in the histology area help me solve the problem of bubbles developing under my coverslips?